Over 90 Total Lots Up For Auction at One Location - WA 04/08

siemens mobilett xp no error no exposure

(65)
(7)

ghazy

siemens mobilett xp no error no exposure

February 20, 2020 09:42

Hello
The mobilett xp gave me error 6 and that was because F7 (5 amp) blown. After replacing this fuse the equipment work properly, complete the ready, green light lit, no error, but when pressing exposure only the collimater light lit and nothing happen, no x ray, noerror, exposure light does not lit. handpiece checked and it was O.K.
any help will be appreciated.

Log inor Register

to rate and post a comment

(17)
(4)

GABRIEL GUTIERREZ

re: siemens mobilett xp no error no exposure

February 21, 2020 10:36

Can you hear the tube anode rotation?

Log inor Register

to rate and post a comment

(65)
(7)

ghazy

re: re: siemens mobilett xp no error no exposure

February 24, 2020 09:22

Yes i can hear the rotating

Log inor Register

to rate and post a comment

(65)
(7)

ghazy

re: re: siemens mobilett xp no error no exposure

February 24, 2020 09:22

yes

Log inor Register

to rate and post a comment

(65)
(7)

ghazy

re: re: siemens mobilett xp no error no exposure

February 24, 2020 09:29

I can hear rotor sound. All voltages exist. I have only led (V33) in (D1) board does not lit. all fuses checked

Log inor Register

to rate and post a comment

(136)
(25)

Germni

re: siemens mobilett xp no error no exposure

February 24, 2020 09:32

ERR 6 Ih < I preheat (filament current error)
System
mode:
Standby
Explanation: The instantaneous filament current (test point IH) is less than
the specified nominal value (test point IHS) of the microcontoller (D916).
Action:
• Check the resonant filament current (D916; TP-I), the filament current
(D916, TP-IH), and the nominal filament current value (D916, TP-IHS). IH
and IHS must be the same. (Further information is available in the "Filament
circuit"
section
of this document).
Possible causes: Defective cable/contacts between D927 (X5) and the single
tank; fuse
F5 blown on D927;
D927; D916;
ribbon
cable
X11
(D927-D916);
single
tank (heating
coil,
filament
transformer).

Log inor Register

to rate and post a comment

(65)
(7)

ghazy

re: re: siemens mobilett xp no error no exposure

February 26, 2020 09:45

Sorry for the mistake. the model is mobilett plus not mobilett xp. i am very sorry. in the errors list error 6 is related to rotation speed.
I checked the connections between single tank and the equipment all were ok. all voltages exist.
If you please have the wiring diagram for this model i will appreciate send me a copy.

Log inor Register

to rate and post a comment

(65)
(7)

ghazy

re: re: siemens mobilett xp no error no exposure

February 26, 2020 09:45

Oh, when i tried to do P12 and P13 tests i could not reach them because the list in screen jumps from P11 to P14 directly

Log inor Register

to rate and post a comment

(17)
(4)

GABRIEL GUTIERREZ

re: re: re: siemens mobilett xp no error no exposure

February 28, 2020 10:23

Which serial number is it?
P13 & P14 are firmware dependent

Log inor Register