david sumpter
seimens mobilete hibrid xp
March 18, 2015 04:03
need service manuals for the mobilete hybrid xp
the system is showing error 35 any coments?
to rate and post a comment
Lenny Place
re: seimens mobilete hibrid xp
March 19, 2015 01:41
Send me an email address and I will send you the manual
ERR 35 Short-circuit at the main inverter
System mode:
Exposure Explanation: D916 (CPU) detects a main inverter short-circuit.
Action:
· Check test point TP - SCM on D916 (CPU) after exposure, (high = error).
· ERR 35 with exposures in all kV ranges indicates a probable single tank defect.
· Check the cable connections between D916 and D962 (X20).
· Check the cable connections between D962 and the single tank (U - V).
· Check C1.
· Check the grounding cable connection between D972 and the device frame (on the voltage supply side U1/U2).
to rate and post a comment
(2)
david sumpter
re: re: seimens mobilete hibrid xp
March 19, 2015 10:46
dssl60@icloud.com thanks leny
to rate and post a comment
Lenny Place
re: re: re: seimens mobilete hibrid xp
March 19, 2015 04:45
just sent you a link
to rate and post a comment
(19)
Mikhail Zakurdaew
re: re: re: re: seimens mobilete hibrid xp
April 10, 2015 09:19
Hello!
Lenny, please send me a copy of the service manual for the Mobilett XP Digital.
Thanks in advance, with respect.
Mikhail Z. mhali@mail.ru
to rate and post a comment
Lenny Place
re: re: re: re: re: seimens mobilete hibrid xp
April 10, 2015 03:40
look for an email Dropbox
to rate and post a comment
Alma Celis
re: re: re: re: re: re: seimens mobilete hibrid xp
December 14, 2017 11:58
hello, someone will have the selectors k1 and k2 for a siemens mobilete xp hybrid
to rate and post a comment