hi, everyone.
I have problem with Siemens Somatom Spirit. Sometimes when start exam make topogram and on 3-4 axial scan system stop scan without any error on display . Wait 10-15 second and start again. On eventlog I found error CT_GPC_E_10_42_P1P2
Thanks
GPC means that you have a problem with Gantry Panel Control, it may opens the stop loop. Also sometimes this problems happens when the Sprocket coupled to the gantry that counts its rotation has some Teeths missed.
(2)
Georgi Gaydov
Problem with CT Siemens Somatom Spirit
February 08, 2019 09:08
hi, everyone.
I have problem with Siemens Somatom Spirit. Sometimes when start exam make topogram and on 3-4 axial scan system stop scan without any error on display . Wait 10-15 second and start again. On eventlog I found error CT_GPC_E_10_42_P1P2
Thanks
to rate and post a comment
(4)
MIGUEL SAYRITUPAC
re: Problem with CT Siemens Somatom Spirit
February 10, 2019 01:30
Hello George.
Write to me miguel.araujo2270@gmail.com.
To atach pic of event log.
Best regards.
Miguel Angel
to rate and post a comment
(67)
Gabriel Cool
re: Problem with CT Siemens Somatom Spirit
February 10, 2019 01:31
Could you send me photo of error on email?
to rate and post a comment
(1)
Hesham Rihan
re: Problem with CT Siemens Somatom Spirit
February 10, 2019 01:33
can you send me on hesham.rihan@imagemedical.net a snapshot of the event log or even an extract of it and i will try to help
to rate and post a comment
Juan Jose Garate Pirela
re: Problem with CT Siemens Somatom Spirit
February 15, 2019 09:11
GPC means that you have a problem with Gantry Panel Control, it may opens the stop loop. Also sometimes this problems happens when the Sprocket coupled to the gantry that counts its rotation has some Teeths missed.
to rate and post a comment
(2)
Georgi Gaydov
re: re: Problem with CT Siemens Somatom Spirit
March 12, 2019 01:19
Hi!
Sorry for may late reply.
The problem is on anode rotation of x ray tube.
The tube have very big noise and vibration when turn on with 60HZ/150HZ and sometimes the tilt sensor is changing to +1;0;-1 and stop scan.
to rate and post a comment