Philips microdose L15 error 11-4-2

Tags: Mammo, PHILIPS

(3)
(1)

Luis Fernandez

Philips microdose L15 error 11-4-2

October 12, 2021 10:05

Can someone give me information to solve a philips microdose L150 mammograph that gives me an error 11-4-2?

Log inor Register

to rate and post a comment

(4)

Leduan GONZALEZ

re: Philips microdose L15 error 11-4-2

October 13, 2021 10:34

Error 11-4-2 microdose philips.
Timeout resetting HVG. Please reset. The HVG did not respond to a reset request.
1. Try resetting again.
2. Restart the HVG (or entire system) and try again.
3. Check cabling between SCS and HVG.
4. Replace HVG.

Log inor Register

to rate and post a comment

(13)

EDUARDO BUSTOS

re: Philips microdose L15 error 11-4-2

October 13, 2021 10:38

11-4-2
Timeout resetting HVG. Please reset.
The HVG did not respond to a reset request.
1. Try resetting again.
2. Restart the HVG (or entire system) and try again.
3. Check cabling between SCS and HVG.
4. Replace HVG.