Luis Fernandez
Philips microdose L15 error 11-4-2
October 12, 2021 10:05
Can someone give me information to solve a philips microdose L150 mammograph that gives me an error 11-4-2?
to rate and post a comment
Leduan GONZALEZ
re: Philips microdose L15 error 11-4-2
October 13, 2021 10:34
Error 11-4-2 microdose philips.
Timeout resetting HVG. Please reset. The HVG did not respond to a reset request.
1. Try resetting again.
2. Restart the HVG (or entire system) and try again.
3. Check cabling between SCS and HVG.
4. Replace HVG.
to rate and post a comment
EDUARDO BUSTOS
re: Philips microdose L15 error 11-4-2
October 13, 2021 10:38
11-4-2
Timeout resetting HVG. Please reset.
The HVG did not respond to a reset request.
1. Try resetting again.
2. Restart the HVG (or entire system) and try again.
3. Check cabling between SCS and HVG.
4. Replace HVG.
to rate and post a comment