Philips BV Endura error M325

Tags: C-Arm

Posted by an Anonymous User

Philips BV Endura error M325

June 17, 2019 05:11

Can anyone help me please with the error M325

Log inor Register

to rate and post a comment

(22)
(6)

Sergey Radionov

re: Philips BV Endura error M325

June 19, 2019 09:19

X-ray generator not available. Switch system off/on. Call service if problem is persistent.
Logfile text: The self test of the FA Generator Control Board SHA4 has failed. The X-ray function is disabled.
Check the cabling between the Generator Control Board and the SE unit. Replace the Generator Control Board.