Javier Aguilera

Norland Eclipse bone densitometer High Atten. Filter Failure.

August 19, 2014 05:41

Hello everyone, I have a daily calibration problem with a Norland Eclipse Bone densitometer. While doing the daily verification/calibration, the unit fails to pass the High Atten. Filter Test. I would appreciate any help on solving this issue.

Thank you in advance.

1 person liked this
Log inor Register

to rate and post a comment

(192)
(44)
(1)

Dante Nuevo

re: Norland Eclipse bone densitometer High Atten. Filter Failure.

August 20, 2014 03:18

Hello Javier.

If the Filters look white or yellow are oxidized and damaged.
I had the same problem with one and had to change the filters to solve it.

Regards.

Dante Nuevo.

dantenuevo@hotmail.com

Log inor Register

to rate and post a comment

(41)
(6)

Javier Aguilera

re: re: Norland Eclipse bone densitometer High Atten. Filter Failure.

August 21, 2014 12:44

Thanks Dante, I don't see any problem (electrical or mechanical), they just don't pass the attenuation test. I will change them and see what happens.
Best regards,

Log inor Register

to rate and post a comment

(2)

DUFAUT Jack

re: Norland Eclipse bone densitometer High Atten. Filter Failure.

August 21, 2014 12:43

Hello ,
you must make an initial calibration. In 50% of cases it is enough.

Log inor Register