Norland Eclipse bone densitometer High Atten. Filter Failure.
August 19, 2014 05:41
Hello everyone, I have a daily calibration problem with a Norland Eclipse Bone densitometer. While doing the daily verification/calibration, the unit fails to pass the High Atten. Filter Test. I would appreciate any help on solving this issue.
re: re: Norland Eclipse bone densitometer High Atten. Filter Failure.
August 21, 2014 12:44
Thanks Dante, I don't see any problem (electrical or mechanical), they just don't pass the attenuation test. I will change them and see what happens.
Best regards,
GBS offers service, repair, rentals & spare parts for your ventilators. With over 40 years of experience our certified & factory-trained biomedical technicians refurbish all our used ventilators. Call today 800.558.9449
(6)
Javier Aguilera
Norland Eclipse bone densitometer High Atten. Filter Failure.
August 19, 2014 05:41
Hello everyone, I have a daily calibration problem with a Norland Eclipse Bone densitometer. While doing the daily verification/calibration, the unit fails to pass the High Atten. Filter Test. I would appreciate any help on solving this issue.
Thank you in advance.
to rate and post a comment
(44)
(1)
Dante Nuevo
re: Norland Eclipse bone densitometer High Atten. Filter Failure.
August 20, 2014 03:18
Hello Javier.
If the Filters look white or yellow are oxidized and damaged.
I had the same problem with one and had to change the filters to solve it.
Regards.
Dante Nuevo.
dantenuevo@hotmail.com
to rate and post a comment
(6)
Javier Aguilera
re: re: Norland Eclipse bone densitometer High Atten. Filter Failure.
August 21, 2014 12:44
Thanks Dante, I don't see any problem (electrical or mechanical), they just don't pass the attenuation test. I will change them and see what happens.
Best regards,
to rate and post a comment
DUFAUT Jack
re: Norland Eclipse bone densitometer High Atten. Filter Failure.
August 21, 2014 12:43
Hello ,
you must make an initial calibration. In 50% of cases it is enough.
to rate and post a comment