FCR Prima T and console advance software

Tags: PACS / DR / CR / Digital

(2)
(1)

Michał Lysko

FCR Prima T and console advance software

March 31, 2020 10:45

Hello,

There's little problem with console advance software. After scanning plate image isn't automatically transferred to PACS server. It only transfers when technican does it manually. Where can i turn it so images are transferred automaticcaly to PACS ?

Best regards,
Michał

1 person liked this
Log inor Register

to rate and post a comment

(41)
(7)

xtekvet

re: FCR Prima T and console advance software

April 01, 2020 10:58

make sure the have the following 2 things checked in property settings
1. under setup during a study 1 tap click on finish exposure and QA
2.under distribution setting click on distributes all attributes individually.
Then the study will be forwarded to pacs after it is closed.

Log inor Register

to rate and post a comment

(4)

Shawn Broussard

re: FCR Prima T and console advance software

April 02, 2020 11:56

Well in the service utility make sure you have the PACS selected in the distribution code and save your changes before you get out . also to make it automatically send to PACS log into the user utility using admin privileges, either six ones if it’s the old IIP software, or VIP if it’s the newer software, go to parameter batch updating, then in the distribution setting drop-down arrow select your PACS and make sure the little box is checked. That’s it. Let me know if this helps