AGFA CR85-X

(13)
(3)
(6)

Mike Rodemers

AGFA CR85-X

August 20, 2013 04:42

Does anyone know what error code 226C4 means upon power up of CR 85-X. Please advise or if you have service manual with error codes for this unit

Log inor Register

to rate and post a comment

(58)
(16)

Rebel Cook

re: AGFA CR85-X

August 20, 2013 07:10

Power supply test of stepper node belt failed. Check the cables and connectors, fuses, power supply, safety interlocks.

Log inor Register

to rate and post a comment

(54)
(7)
(6)

Luis Diaz

re: AGFA CR85-X

August 20, 2013 07:10

Hello Mike,
The next explications is referent your error in the CR85-X...

226C4
E_IBF_TST_SNBELT_POWER
Reason : Input buffer module: Power supply test of stepper node Belt failed.
Meaning : No Power supply on stepper node detected.
- power supply not connected.
- fuse blown.
- power supply defective.
Cure : - check cables and connectors.
- check/replace fuse.
- check/replace power supply.

Good luck!
Bye

1 person liked this