AGFA CR85-X
August 20, 2013 04:42
Does anyone know what error code 226C4 means upon power up of CR 85-X. Please advise or if you have service manual with error codes for this unit
to rate and post a comment
(16)
Rebel Cook
re: AGFA CR85-X
August 20, 2013 07:10
Power supply test of stepper node belt failed. Check the cables and connectors, fuses, power supply, safety interlocks.
to rate and post a comment
(7)
(6)
Luis Diaz
re: AGFA CR85-X
August 20, 2013 07:10
Hello Mike,
The next explications is referent your error in the CR85-X...
226C4
E_IBF_TST_SNBELT_POWER
Reason : Input buffer module: Power supply test of stepper node Belt failed.
Meaning : No Power supply on stepper node detected.
- power supply not connected.
- fuse blown.
- power supply defective.
Cure : - check cables and connectors.
- check/replace fuse.
- check/replace power supply.
Good luck!
Bye